At what point on the TDS meter....
do you change your filter material on your RO unit? Just wondering.....
Remember when it was at dead ZERO, now about 4 or 5. Seems like nothing, but has me wondering. |
09/08/2007, 02:39 PM | #2 |
.Registered Member
Join Date: Sep 2006
Location: Cincinnati, Ohio
Posts: 4,857
|
Your prefilters don't affect tds levels, so think only about your membrane and DI resin.
When membranes go bad you'll see a reduction in the amount of purified water they produce, or an increase in the tds of your RO water. Replace your DI resin when the DI water climbs to about 2 or 3 ppm. Russ @ BFS |
|
|