I just calibrated it with pinpoints 53.0mS calibration fluid and it was .004 higher than it supposed to be! Then I put some distilled water on it to see what comes up and its .004 below 0. I don't know what to believe now. Any assistance? |
![]() |
![]() |
![]() |
#2 |
Registered Member
Join Date: Jun 2008
Location: Aggieland
Posts: 169
|
RTFM?
__________________
Do you like to 'Do It Yourself'? Current Tank Info: No Current Tanks. I just like their light fixtures. |
![]() |
![]() |
![]() |
#3 |
Registered Member
Join Date: Jul 2004
Posts: 2,356
|
Believe the calibration fluid.
|
![]() |
![]() |
![]() |
#4 |
Registered Member
![]() Join Date: Jul 2007
Location: Baltimore, MD
Posts: 2,664
|
is your refractometer adjustable?? No use in calibrating it if you can't adjust it. If it has a screw, you need to crank it down with the cal fluid until it reads 0.
|
![]() |
![]() |
![]() |
#5 |
Registered Member
Join Date: Mar 2006
Location: West Palm, FL
Posts: 990
|
Whats RTFM?
abulgin, it is calibratable (adjustable). The calibration fluid I have is for 1.026. If I crank it to 0 then when water is 1.026 it will read 0 |
![]() |
![]() |
![]() |
#6 |
Registered Member
Join Date: Mar 2005
Location: Port Clinton, Oh
Posts: 1,470
|
I would trust the calibration fluid over distilled water. At least that is what I have read.
I set mine to 1.026 or 35ppt with the calibration fluid in it. |
![]() |
![]() |
![]() |
#7 |
Mouth Pipetting
Join Date: Aug 2006
Location: Charlottesville, VA
Posts: 6,278
|
Go with the calibration fluid. The problem is that most hobbyist refractometers are not accurate over the range from 0.000 to 1.026; therefore you should calibrate it near the value you intend to use. Its not unusual for RO/DI water to read below 0.000 once the refractometer has been calibrated to 1.026.
__________________
Josh |
![]() |
![]() |
![]() |
#8 |
Registered Member
![]() Join Date: Jul 2007
Location: Baltimore, MD
Posts: 2,664
|
ah, I misspoke--dont know what I was thinking. When you use the cal fluid, you need to crank the reading up to 35 ppt. Then, it is calibrated. Sorry for confusion.
|
![]() |
![]() |
![]() |
#9 | |
Registered Member
Join Date: Jan 2006
Location: maryland
Posts: 6,923
|
Quote:
Listen to this !!!!!!!!!!!!!!!!!!!!!!!
__________________
I found a way to make a small fortune running a reef tank. Start with a large fortune. Unofficial President of the SEACLONE haters club Current Tank Info: 125 mixed reef 110 lbs LR, 1x250watt XM 20K MH 2x175watt XM 20K MH on Magetics 2X96 watt actinic PC, 220 watt VHO actinic, 30 gallon refugium, closed loop system powered by Sequence Dart MSX 200 skimmer 38 gallon sump, Oceansmotions squirt |
|
![]() |
![]() |
![]() |
#10 |
Registered Member
Join Date: Mar 2006
Location: West Palm, FL
Posts: 990
|
aight, thanks a bunch yall
|
![]() |
![]() |
![]() |
#11 |
Reef Chemist
![]() Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 86,233
|
Just so folks understand better, the primary reason that refractometers can be inaccurate for measuring seawater, aside from potential mismanufacturing, is that they are all intended to measure brine solutions. That is, sodium chloride in water. Since seawater has more in it than sodium chloride, that causes problems.
in short, refractometers measure the refractive index of the water, and then try to relate that back to the salinity or specific gravity. But the relationship between refractive index and these other parameters is different for seawater than for sodium chloride solutions. So the best way around that is to calibrate with a solution that is known to match the refractive index of normal 35 ppt seawater. That can be done with a true seawater like solution (such as the Pinpoint standard) or with a sodium chloride solution that has the same refractive index as 35 ppt seawater. It cannot necessarily be done with any other solutions, such as other 53 mS/cm conductivity standards, since they do not necessarily have the refractive index of seawater. I discuss these in more detail here: Refractometers and Salinity Measurement http://reefkeeping.com/issues/2006-12/rhf/index.php Reef Aquarium Salinity: Homemade Calibration Standards http://reefkeeping.com/issues/2004-06/rhf/index.htm
__________________
Randy Holmes-Farley Current Tank Info: 120 mixed reef |
![]() |
![]() |
![]() |
#12 |
Registered Member
Join Date: Aug 2005
Location: Pasadena, TX
Posts: 385
|
cool thanks, now i need to recalibrate my refractometer because i was doing the same thing with ro/di water and adjusting to 0 ppt.
where can i get a few drops of Pinpoint Standard or 35 ppt seawater?
__________________
120g Reef Tank Oceanic, 2x175watt MH 14k Iwasaki, 1 Ecotech MP40, Reeflo Dart return pump, AquaC EV-120 Eheim 1262 ,AquacontrollerJr, KentMaxximaHI-S, PhosBanReactor150, Geo Kalk Rx. Current Tank Info: working on diy led lighting and finishing a sump refuge relocated outside the house |
![]() |
![]() |
![]() |
#13 |
Reef Chemist
![]() Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 86,233
|
You can make the DIY in the article, which is only highly accurate if you have a good balance, or you can buy the Pinpoint from many vendors.
![]()
__________________
Randy Holmes-Farley Current Tank Info: 120 mixed reef |
![]() |
![]() |
![]() |
#14 |
Premium Member
![]() Join Date: Feb 2007
Location: SJ,CA
Posts: 1,694
|
i have a question thought, in the bottle it says something about 77 degrees, does the bottle have to be 77 when you read it, or will it always be at 33ppt no matter what temp
|
![]() |
![]() |
![]() |
#15 |
Reef Chemist
![]() Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 86,233
|
35 ppt, and if it is an ATC (automatic temperature correction) refractometer, then any temp reasonably close to tank temp is OK, and if the ATC is perfect, then any temp is OK.