Do you use temp adjusted pH calibration?
Was just wondering if you use temperature adjusted pH calibration? I initially calibrated my pH probe when I first got my unit (with this enabled).
I rince my pH probe in RO/DI then place in my 7.00 solution pack in a water bath along with my temperature probe so it reads the baths water temperature. I run the calibration reads something like 7.51 offset... rinse in RO/DI then I place the probe in 10.00 solution (again something like 10.52 offset)... rinse in RO/DI again, go back in 7.00 solution to check... sometimes the probe with read 0.03 to 0.05 points off... Is this normal or did I mess up? I let the probe in each solution for about 5min. Each packet is a PinPoint 1 Use pack... I use the same solutions for the entire calibration process...
__________________
Rich Current Tank Info: DSA 135g (48x24x27) Starphire, Elite Aquatics M2 Sump, Reef Octopus Diablo DCS200-INT, GEO 612 Calcium Reactor + AP Carbon Doser, 2x Radion Pro Gen2, 2x VorTech MP40 |
12/10/2011, 06:48 PM | #2 |
WHO DAT??
Join Date: Sep 2009
Location: Baton Rouge, LA
Posts: 5,709
|
It's probably just your probe. If you figure your temp probe is probably only accurate to 1 degree with a tenth of a degree resolution and your pH only varies by a couple hundredths over that same 1 degree, it doesn't make much of a difference.
__________________
I don't work for Neptune. Please put all questions into forum posts and not PM's. Visit the Neptune Community Forums for the latest Neptune updates - forum.neptunesystems.com |
12/10/2011, 07:14 PM | #3 |
Premium Member
Join Date: Aug 2004
Location: Pottstown, PA
Posts: 866
|
Well, I was just wondering about it reading the 0.50 something offset when calibrating. I assume it's the variation in the probe being 'corrected' out. I was use to PinPoint meters and having to zero them to 7.00 or 10.00 with the adjustment screws...
I clean and calibrate about once a month. I assume thats still OK? How much drift should I expect to see, any idea? FYI, love the "Guide"...
__________________
Rich Current Tank Info: DSA 135g (48x24x27) Starphire, Elite Aquatics M2 Sump, Reef Octopus Diablo DCS200-INT, GEO 612 Calcium Reactor + AP Carbon Doser, 2x Radion Pro Gen2, 2x VorTech MP40 |
12/10/2011, 08:04 PM | #4 | |
WHO DAT??
Join Date: Sep 2009
Location: Baton Rouge, LA
Posts: 5,709
|
Quote:
__________________
I don't work for Neptune. Please put all questions into forum posts and not PM's. Visit the Neptune Community Forums for the latest Neptune updates - forum.neptunesystems.com |
|
03/28/2012, 08:26 AM | #5 |
Certified Reef Fanatic
Join Date: Nov 2011
Location: Peoria, il
Posts: 220
|
The buffers for pH calibration are designed for use at 25c/77f. I run at 80 degrees so I must increase my calibration setpoints .05. Here is a handy table.
pH Temperature Correction Factors at Varying Temperatures from Standard (25°C) Temp°C Correction Factor 40 Plus 0.45 39 Plus 0.42 38 Plus 0.39 37 Plus 0.36 36 Plus 0.33 35 Plus 0.30 34 Plus 0.27 33 Plus 0.24 32 Plus 0.21 31 Plus 0.18 30 Plus 0.15 29 Plus 0.12 28 Plus 0.09 27 Plus 0.06 26 Plus 0.03 25 0.00 24 Minus 0.03 23 Minus 0.06 22 Minus 0.09 21 Minus 0.12 20 Minus 0.15 19 Minus 0.18 18 Minus 0.21 17 Minus 0.24 16 Minus 0.27 15 Minus 0.30 14 Minus 0.33 13 Minus 0.36 12 Minus 0.39 11 Minus 0.42 10 Minus 0.45 |
Tags |
calibration, probe |
|
|
Similar Threads | ||||
Thread | Thread Starter | Forum | Replies | Last Post |
Adjusting PH Calibration Data? | zhangg3 | GHL /ProfiLux / Mitras | 5 | 02/28/2011 10:02 AM |
need help Calibrating Ph Probe | plankton99 | GHL /ProfiLux / Mitras | 5 | 02/13/2011 09:46 PM |
pH Calibration -- manual adjustments? | tgreene | GHL /ProfiLux / Mitras | 14 | 09/29/2010 11:11 AM |
Temp Adjustment on pH | alrha | Neptune Systems | 4 | 10/30/2009 08:04 AM |
new setup, wrong temp, pH - calibrate or not? | sjp770 | GHL /ProfiLux / Mitras | 8 | 12/17/2007 09:04 PM |