Hanna phosphate checker error of 1nu
The last two days after the checker gets to zero it gives the error code of 1nu. I looked at manual and it states "the sample and the zero cuvette are inverted". What does this mean? I am not doing anything different than before that I can think of. Thanks for the input.
|
04/11/2013, 02:55 PM | #2 |
Registered Member
Join Date: Sep 2005
Location: RI
Posts: 1,085
|
You can basically read it as zero. The actual code is "INV" and not "1nu." It means that the calibration sample looked darker than the test sample.
|
|
|