11/05/2016, 10:46 AM | #1 |
Registered Member
Join Date: Aug 2014
Posts: 1,432
|
2 mL Sample for Hanna P ULR
I needed to perform a number of tests on small aquarium water samples. Recently, I found a way to use a 2 instead of a 10 mL sample in the Hanna ULR Phosphorous Checker. The test vial needed to be raised about 24 mm which was accomplished with a five mL syringe barrel cut down with an Xacto saw and placed in the Checker sample compartment. Because the test vial was now too high for the Checker lid to close, an opaque medicine bottle served as a cover during the test. For the normal 10 mL sample, 0.16 mL of powder reagent is used. Using a 1 mL syringe with the end cut off just below the "0" mark, 0.03 mL, a little less than 1/5 the amount used for a 10 mL sample, was dosed to the sample. The plot below of the Hanna reading versus the phosphate standard validates the use of a 2 mL sample.
|
11/05/2016, 09:52 PM | #2 |
RC Mod
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
|
Hmm, most of those points are quite close. I wonder what happened at around 100 ppm P-PO4. That's not bad for the cost!
__________________
Jonathan Bertoni |
11/06/2016, 08:16 AM | #3 | |
Registered Member
Join Date: Aug 2014
Posts: 1,432
|
Quote:
|
|
11/06/2016, 03:54 PM | #4 |
RC Mod
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
|
I agree that checking the sensitivity to the quantity of reagent would be a good next step.
__________________
Jonathan Bertoni |
01/29/2017, 05:56 AM | #5 |
Registered Member
Join Date: Aug 2014
Posts: 1,432
|
A quick update. I did extensive testing of the 2 mL sample protocol and found it gave variable results and PO4 values much lower than the same sample tested at 10 mL. I don't know what's going on, so, I am suspending use of this approach until I can solve these issues. Sorry about the misleading information.
Dan |
01/29/2017, 09:28 PM | #6 |
RC Mod
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
|
Oh, well, it's an interesting idea. It might require a precision scale to get it working, though.
__________________
Jonathan Bertoni |
01/30/2017, 05:22 PM | #7 |
Registered Member
Join Date: Aug 2014
Posts: 1,432
|
|