Calibrating PH Probe
I calibrated my PH probe using Milwaukee PH 7.01 & 10.01 buffer solutions.
When I put the probe in the 7.01 solution it stabilized at 7.76 In the 10.01 solution it stabilized at 10.99. Are these numbers accurate? Or should the PH numbers been closer to 7.01 & 10.01 when I put the probe into buffers?? After putting the probe in the tank, it shows my PH is 8.03, and tested at 8.20 with a test kit. |
10/03/2005, 09:56 AM | #2 |
RC Sponsor
Join Date: Aug 2001
Location: San Jose, CA
Posts: 4,311
|
The numbers displayed during calibration don't matter. After calibration with the controller in the run display, double check that your probe has been properly calibrated, by putting it in the calibration solutions.
Curt |
10/03/2005, 06:47 PM | #3 |
Registered Member
Join Date: Feb 2004
Location: Ohio
Posts: 82
|
Good idea. Why didn't I think of that..... Thanks for your help!
|