RO/DI troubleshooting help
I've got a 4 stage RO/DI filter with a DM1 inline TDS monitor.
Incoming tapwater is 340ppm DI output is 9ppm RO output is 16ppm So based on my calculations the RO is rejecting about 95%. I don't have the specs. on the RO membrane but i believe this should be fine. So i replaced the filter block and carbon block and the DI block and the DI output dropped down to 2 PPM. So my question is.....what is the problem? Why can't i get it down to 0? I have an 800 PPM calibration solution and based on the temp/ppm chart on the solution the TDS monitor appears to be reading just fine. I'm baffled. Any ideas?
__________________
June 2001 Tank of the Month. Yes, i said 2001. Current Tank Info: 112G 36"x36"x20" Lee-mar rimless. Build Thread: http://www.reefcentral.com/forums/showthread.php?t=2293182. |
11/28/2008, 09:58 PM | #2 |
Registered Member
Join Date: Nov 2008
Location: Charleston S.C.
Posts: 771
|
Do you have at least 65 water pressure. If not it could be dropping by the time it reaches your last filter. . That could cause your tds to read higher.
|
11/28/2008, 10:48 PM | #3 | |
Team RC Member
|
Quote:
I think i found my problem. I tried switching the probes on my TDS monitor and put the input probe on the output line and it reads zero now. Looks like the output probe may be bad. Unfortunately the TDS monitor probes can't be adjusted individually so its most likely toast. I'll be buying the DM2 model where the probes can be adjusted individually. Lesson learned.