Not sure if i call it chemistry, but Hydrometer issue

I set up a new biocube 32, and have 2 hydrometers. one is instant ocean and is reading 1.030. the other is a cheap floating type and reads 1.022 with filter and power head off.

When put in D.I. water cheap one is right on 1.000, instant ocean one is 1.004

we have a lab where I work, and I plan to weigh the water tomorrow and get a real number. I just don't want to do this all the time.

Do I really need to purchase a refractometer (sp) ? From what I have read they are kind of difficult to calibrate.

Or use another type of hydrometer.

I did clear the bubbles of the instant ocean hydrometer needle.


davered00ss is offline   Reply With Quote
Unread 01/31/2017, 06:13 PM   #2
mcgyvr
Registered Member
 
mcgyvr's Avatar
 
Join Date: Sep 2003
Location: North Carolina
Posts: 20,050
well.. you really don't need a refractometer once you can verify your hydrometers accuracy.. They seem to stay consistent forever/long time. (be it consistently high or low,etc..)
But a refractometer isn't hard to calibrate either and does add a little extra piece of mind..
I just drop a few drops of the calibration solution on it and turn a screw as needed to calibrate it..


__________________
Who me?
mcgyvr is offline   Reply With Quote
Unread 01/31/2017, 06:32 PM   #3
cnmcalpi
Moved On
 
Join Date: Sep 2010
Posts: 46
you can use randys calibration fluid to find out where its inaccuracy lies. Mine reads about .001 low. 1.025 is actually 1.026.


cnmcalpi is offline   Reply With Quote
Unread 01/31/2017, 06:35 PM   #4
bertoni
RC Mod
 
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
You probably can calibrate the swing-arm with a saltwater standard, like the PinPoint Salinity Calibration Fluid. Personally, I went with a conductivity meter since it's a bit more flexible than a refractometer, but either of those types of devices probably would be easier to use. They are easy to calibrate, in my experience, as well.


__________________
Jonathan Bertoni
bertoni is offline   Reply With Quote
Unread 01/31/2017, 06:43 PM   #5
cnmcalpi
Moved On
 
Join Date: Sep 2010
Posts: 46
The calibration fluid would be lab grade. +1


cnmcalpi is offline   Reply With Quote
Unread 01/31/2017, 06:49 PM   #6
timnem70
Registered Member
 
timnem70's Avatar
 
Join Date: Apr 2015
Posts: 462
Throw it away and the 30 bux. Its just another Better way of reading.However, I used one for years and like mcgyvr says was a longtime standard. It's just that if you even get a micro bubble under the swing arm you're looking at the wrong gravity. As I said one reads specific gravity and the refractometer determines displacement, a hydrometer density.

Sent from my SM-G930V using Tapatalk


timnem70 is offline   Reply With Quote
Unread 01/31/2017, 06:49 PM   #7
timnem70
Registered Member
 
timnem70's Avatar
 
Join Date: Apr 2015
Posts: 462
And spend 30 bux.

Sent from my SM-G930V using Tapatalk


timnem70 is offline   Reply With Quote
Unread 01/31/2017, 07:05 PM   #8
cnmcalpi
Moved On
 
Join Date: Sep 2010
Posts: 46
I have a little plastic piece that I stick in my swing arm to bang it around and dislodge it/knock off any bubbles. After using it for years and knowing the calibration, Im fairly sure that salinity is spot on.

I would like to have a refractometer, though. Do they get knocked out of adjustment and need calibration often?


cnmcalpi is offline   Reply With Quote
Unread 01/31/2017, 07:32 PM   #9
davered00ss
Registered Member
 
davered00ss's Avatar
 
Join Date: Jan 2017
Location: Canterbury CT
Posts: 84
Quote:
Originally Posted by bertoni View Post
You probably can calibrate the swing-arm with a saltwater standard, like the PinPoint Salinity Calibration Fluid. Personally, I went with a conductivity meter since it's a bit more flexible than a refractometer, but either of those types of devices probably would be easier to use. They are easy to calibrate, in my experience, as well.
What reading are you looking for on the conductivity meter? I have a Hach portable one in my office at work that is not in use lol and calibration standards for it.


davered00ss is offline   Reply With Quote
Unread 01/31/2017, 07:37 PM   #10
davered00ss
Registered Member
 
davered00ss's Avatar
 
Join Date: Jan 2017
Location: Canterbury CT
Posts: 84
I manage a wastewater treatment plant for a small city. We have a pretty decent lab on site to test water and wastewater. its easy to bring in water to work and test it, or borrow equipment overnight


davered00ss is offline   Reply With Quote
Unread 01/31/2017, 07:46 PM   #11
bertoni
RC Mod
 
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
Saltwater is about 53.0 mS/cm at 75 F.


__________________
Jonathan Bertoni
bertoni is offline   Reply With Quote
Unread 02/01/2017, 05:03 AM   #12
Ron Reefman
Registered Member
 
Ron Reefman's Avatar
 
Join Date: Jul 2009
Location: Cape Coral, FL
Posts: 10,431
Take in a sample of your water and your hydrometer. Test the water with your lab equipment. Lets just say it tests at 1.026. Then test the water with your hydrometer and lets say it reads 1.030. Then you'll know the hydrometer reads 0.004 too high. Mark your hydrometer (I put a label on mine) so you know the hydrometer reads 0.004 too high. I made the label say, "subtract 0.004". If you feel this isn't enough, the next time the salinity in your tank is different, test it like this again. I'll bet you'll find the hydrometer is off by the same 0.004 as it was in the first test.

I have a refractometer and calibration fluid as well as 2 hydrometers. I did this test about 5 years ago and did the retest every 3 months for a year and then twice a year for the next 2 years. I recalibrated the refractometer each time. BTW, sometimes it needed the calibration and sometimes it didn't. But both hydrometers had the very same error rate they had in the first test. Never failed, always the same.

Personally I find the hydrometer easier to use and now I never consider if it is right or wrong. I know it's right. And if I test with the refractometer and it's different than the hydrometer, then I test it with the calibration fluid and guess what? I find that it's out of calibration and the cheap hydrometers were right!


__________________
The good thing about science is that it's true whether or not you believe in it. (Neil deGrasse Tyson)
Visit my build thread http://www.reefcentral.com/forums/showthread.php?t=2593017
Ron Reefman is offline   Reply With Quote
Unread 02/01/2017, 05:40 AM   #13
homer1475
Registered Member
 
homer1475's Avatar
 
Join Date: Mar 2015
Location: Upstate NY
Posts: 5,313
Although I check my refractometer nearly every time I use it, I rarely find it off. I actually only ever have to calibrate it whenever I drop it, or it gets a hard bump. It'sa typically spot on 99% of the time.

I used to use a hydrometer too, there is just way to many variables at play for them to be as accurate as a properly calibrated refractometer.

Or as Ron suggested, use one to calibrate the other.


__________________
80G SCA Build: http://reefcentral.com/forums/showthread.php?t=2560256

Originally posted by der_wille_zur_macht:

"He's just taking his lunch to work"
homer1475 is offline   Reply With Quote
Unread 02/01/2017, 06:58 AM   #14
davered00ss
Registered Member
 
davered00ss's Avatar
 
Join Date: Jan 2017
Location: Canterbury CT
Posts: 84
Looks like the Instant Ocean Hydrometer was right. After weighing 100 ml of water, I came up with a specific gravity of 1.0327

Looks like I need to add some D.I. Water to the tank.


davered00ss is offline   Reply With Quote
Unread 02/01/2017, 07:12 AM   #15
Elricsfate
Registered Member
 
Elricsfate's Avatar
 
Join Date: Oct 2016
Location: Fredericksburg, Va
Posts: 482
I use the Milwaukee digital. No fuss, no muss.


Elricsfate is offline   Reply With Quote
Unread 02/01/2017, 12:03 PM   #16
ReefsandGeeks
Registered Member
 
ReefsandGeeks's Avatar
 
Join Date: Sep 2014
Location: Pittsburgh, PA
Posts: 1,708
Be careful when measuring the volume of the water to check the SG. you need to be able to measure the volume very accurately in order to determine this, and most graduated cylinders are +-5%. I even just got a set of nice calibrated ones used to calibrate other cylinders and they aren't even accurate enough to determine SG to the level I would want.

This can be done with the right equipment though. I have an extreamaly sensitive scale in my lab (resolution 0.00mg) and I use it with a graduated cylinder to weight out 100g of DI water, then note exactly what the cylinder says the volume is. Then I empty and dry the cylinder, and fill it to the same exact volume I recorded before and weigh again. This provides an accurate SG for the fluid.

In order to measure to 1.0264 SG +-0.0001 you need all measurments to have less than 0.01% error cumulatively. Just keep that in mind if that's a route you want to take.

I just recently switched from my swing arm to a refractometer. I'm very happy with the switch as I have more confidence in the repeatability of the refractometer, even if you do need to check it's calibration frequently, it's very easy and quick to do, and no worries about micro bubbles or anything like that.


ReefsandGeeks is offline   Reply With Quote
Unread 02/01/2017, 12:19 PM   #17
ericarenee
Registered Member
 
ericarenee's Avatar
 
Join Date: Sep 2013
Location: Louisville Kentucky
Posts: 3,110
Quote:
Originally Posted by Ron Reefman View Post
Take in a sample of your water and your hydrometer. Test the water with your lab equipment. Lets just say it tests at 1.026. Then test the water with your hydrometer and lets say it reads 1.030. Then you'll know the hydrometer reads 0.004 too high. Mark your hydrometer (I put a label on mine) so you know the hydrometer reads 0.004 too high. I made the label say, "subtract 0.004". If you feel this isn't enough, the next time the salinity in your tank is different, test it like this again. I'll bet you'll find the hydrometer is off by the same 0.004 as it was in the first test.

I have a refractometer and calibration fluid as well as 2 hydrometers. I did this test about 5 years ago and did the retest every 3 months for a year and then twice a year for the next 2 years. I recalibrated the refractometer each time. BTW, sometimes it needed the calibration and sometimes it didn't. But both hydrometers had the very same error rate they had in the first test. Never failed, always the same.

Personally I find the hydrometer easier to use and now I never consider if it is right or wrong. I know it's right. And if I test with the refractometer and it's different than the hydrometer, then I test it with the calibration fluid and guess what? I find that it's out of calibration and the cheap hydrometers were right!
i have been saying this all along. Consistence is the most important part , If you use a hydrometer the same each time.. You will get correct results each time.Thou they may a bit off actually scientific exact results..

If you use a refractormeter that you do not keep calibrated you will get results that will very..

the cheaper of the two is much more accurate.. As you have done just confirm where it stands against scientific exact results and be done with it...

the only benefit of using refractometer is .. ohh nevermind...
Thank you.....


__________________
240+G Mixed BB Reef tank.. 350 G Marine Pond. And the expensive stuff that runs it.
Chic's are for Chic's You silly men Go Fishing or something...

Current Tank Info: 240= gal Reef /550 Gallon Saltwater pond 72 G Bay front Tropical aquarium
ericarenee is offline   Reply With Quote
Unread 02/01/2017, 01:12 PM   #18
davered00ss
Registered Member
 
davered00ss's Avatar
 
Join Date: Jan 2017
Location: Canterbury CT
Posts: 84
I might just try to pick up something like this.

http://www.labdepotinc.com/p-61277-h...FUVLDQod7NEA_Q

Then use a 1000 ml graduated cylinder to float it in?


davered00ss is offline   Reply With Quote
Reply


Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is On



All times are GMT -6. The time now is 08:35 AM.


Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Powered by Searchlight © 2024 Axivo Inc.
Use of this web site is subject to the terms and conditions described in the user agreement.
Reef CentralTM Reef Central, LLC. Copyright ©1999-2022
User Alert System provided by Advanced User Tagging v3.3.0 (Pro) - vBulletin Mods & Addons Copyright © 2024 DragonByte Technologies Ltd.