Just Bought A Pinpoint PH, quick user question.
Bought it used with no instruction manual. I got some free calibration fluid with it made by coralife (i know, not the best) a 7.0 solution and 10.0 bottle.
I see two screw adjustments on the face of the pinpoint meter. How exactly do I adjust the PH reading, wich one of the screws does it. And what is the other screw there for? Thanks P.S. Should I throw the 7.0 solution away. I got two bottles of each. Last edited by cham; 08/09/2006 at 03:22 PM. |
08/09/2006, 05:20 PM | #2 |
Registered Member
Join Date: Jun 2006
Location: Winter Garden, FL
Posts: 3,632
|
No one has a pinpoint PH monitor?
|
08/09/2006, 05:24 PM | #3 |
Registered Member
Join Date: Jun 2005
Location: palm coast, fl.
Posts: 216
|
pinpoint
1 screw is for 7 cal. fluid/ other is for 10 cal.fluid they should be marked respectfully
__________________
STEVE Current Tank Info: 65 gal. softie tank/ new 120 sps in the works, 40 gal sump/ mag 18 return/ 1/4 delta star/ 25watt uv/ 2x250 ARO 10k ushio's/ 2 - 110 watt reef suns/ 2- 110watt actinics /DELTEC-AP 600 skimmer/ 150 gal. cichlid tank |
08/09/2006, 05:28 PM | #4 |
Registered Member
Join Date: Jun 2006
Location: Winter Garden, FL
Posts: 3,632
|
Just noticed that. So its a simple as putting the probe in the fluid and turning the screw to match the fluid you have it in.
Should I use the 7 fluid or 10 or both? Do I need to pour out the solution that I use to calibrate after each test or can I reuse the fluid? |
08/09/2006, 05:38 PM | #5 |
Registered Member
Join Date: Nov 2003
Location: Rhode Island
Posts: 453
|
For my PinPoint I use the 7 and the 10. I buy the little PinPoint packs of cal fluid, just cut them open and throw them out when finished. I would probably not re-use the cal fluid.
|
08/09/2006, 05:39 PM | #6 |
Registered Member
Join Date: Dec 2004
Location: Pawleys Island, SC
Posts: 295
|
No, it's not that simple. You want to use both solutions. One screw is used for the 7.0 solution and the other is used for the 10.0.
You need to rinse the probe between solutions(using tap water) and allow the reading to stabilize before adjusting. You will probably need to repeat the adjustment 3 times. You are done when the readings settle in at 7.0 and 10.0 without an adjustment. I am not familiar with the solutions you have but I use the small foil package solutions and throw them out when done. |
08/09/2006, 06:50 PM | #7 |
Registered Member
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 12,010
|
You should also have a jar of clean water on hand to rinse after each test.
7.0 - adjust - rinse in clean water 10.0 - adjust - rinse in clean water Repeat until no more adjustment are necessary. As stated above it may take 3 or 4 times to get it stabilized. |
08/09/2006, 07:09 PM | #8 |
Registered Member
Join Date: Mar 2006
Location: SC
Posts: 777
|
|
08/09/2006, 07:20 PM | #9 |
Registered Member
Join Date: Jun 2006
Location: Winter Garden, FL
Posts: 3,632
|
Thanks so much for the responces. Boomer helped me out in another thread and it turns out my probe is no good. I have not yet paid the guy for this (brought it home to see if it would work first) and after finding out a new probe is $50 to my door, or I could have a whole new set up for $80 shipped to my door it looks like getting a new one is a no brainer.
|
08/09/2006, 09:20 PM | #10 |
Registered Member
Join Date: Apr 2003
Location: Lake Charles, LA
Posts: 304
|
With calibration fluid, once you open the container and expose the sample to air, the pH will begin to change. It may not be accurate to use after the first use. Better to go with the single use pouches and throw it away when you are done. Always use fresh fluid to calibrate your meter, or the results you get may be inaccurate.
|
|
|